Inter-lab variability in peptide testing — July 2024

Dr.GastroMayo Fri, Sep 19, 2025 at 1:04 AM 41 replies 1,493 viewsPage 1 of 9
This thread is more than 8 months old. Information may be outdated. Consider searching for more recent discussions.
Dr.GastroMayo
VIP Member
2,345
13,456
Jan 2024
Mayo Clinic, MN
Sep 19, 2025 at 1:04 AM#1

Two reports on the same product from two laboratories, four percentage points apart, and no method stated on either.

The bit I cannot resolve on my own is whether a purity figure and a content figure can disagree without one of them being wrong.

I would rather have one careful answer than five confident ones.

49 19mike_nyc, VendorMark, COA_Karl and 46 others
Reply Quote Save Share Report
Dr.AddMedPHL
Senior Member
1,234
6,234
Mar 2024
Philadelphia, PA
Sep 19, 2025 at 2:40 AM#2

Answering the narrow version, because the broad one does not have a single answer. The provenance questions are more useful than the numbers. Does the report name the batch on your vial. Is the date consistent with when that batch existed. Is it addressed to the vendor or to the person who paid for the test. Was the sample taken from general stock or supplied by the vendor. A perfect certificate for a different batch is worth nothing, and vendor-supplied samples are the most common way a genuine report ends up misleading.

Last edited: Sep 19, 2025 at 4:40 AM
48 18sarah_nash92, FitDadDave, RunnerRach and 45 others
Reply Quote Save Share Report
RegAffairsDC
Member
678
3,456
May 2024
Washington, DC
Sep 19, 2025 at 4:16 AM#3
Dr.AddMedPHL said:
The provenance questions are more useful than the numbers.

Agreed, and mass spec answers a question purity does not: identity. High purity tells you one thing dominates the chromatogram, not that it is the molecule you ordered. A mass within about a dalton of expected is the identity check.

Happy to go further on any of that.

47 17hannah_MT, Dr.SportsMedIN, amy_econ_NJ and 44 others
Reply Quote Save Share Report

Janoshik Analytical — Independent Testing

Trusted third-party HPLC & mass spectrometry analysis. Verify peptide purity with the lab the community relies on. Independent. Accurate. Transparent.

Verify Your Peptides

GL Biochem (Shanghai) Ltd. — Direct Manufacturer

Est. 1998. The synthesis house behind the vials you send for testing. ISO 9001 and cGMP certified, 1,500+ staff, batch-specific COA with every order.

Browse GL Biochem
carl_compliance
Member
234
1,123
Nov 2024
Raleigh, NC
Sep 19, 2025 at 5:52 AM#4
Dr.GastroMayo said:
Two reports on the same product from two laboratories, four percentage points apart, and no method stated on either.

Can confirm the pattern Dr.GastroMayo describes. Inter-laboratory variation comes from real methodological differences, not sloppiness: which reference standard was used and how it was itself calibrated, the gradient and column, the detection wavelength, and where the integration baseline was drawn. A percentage point or two between competent labs is expected. What is not expected is a large gap, and the usual explanation there is that one report is not of the batch in your hand.

46 16FDA_TrackerJim, ricardo_MIA, BrianDallas92 and 43 others
Reply Quote Save Share Report
Dr.CardioMD
VIP Member
2,678
14,567
Dec 2023
Cleveland, OH
Sep 19, 2025 at 3:13 PM#5

From the other side of the consultation, briefly.

Vendor quality matters when discussing certificates of analysis. I've tested my last 6 batches through Janoshik and purity has ranged from 99.1% to 100.7%. Consistency is key.

For anyone new: always request the COA, and consider independent verification. The $300 for a Janoshik test is worth the peace of mind.

45 15josh_phd_bmore, roxy_nash, tony_orlando and 42 others
Reply Quote Save Share Report
1239

Similar Threads

HPLC vs UPLC for peptide purity — method comparison study18 replies
Mass spectrometry for peptide identity verification — ESI-MS guide7 replies
Endotoxin testing methods — LAL vs recombinant Factor C17 replies
Red flags on COAs — how to spot a fake certificate5 replies
USP reference standards for peptide verification — sourcing guide3 replies
ForumsNewTrendingMembersAccount

Log In

Forgot password?
No account? Register