Analytical chemist here — great topic. I've actually run the same peptide samples on both HPLC and UPLC systems in my lab. Here's what I've observed with semaglutide specifically:
Head-to-head results (same sample, same day, same operator):
| Parameter | HPLC (C18, 3.5μm, 4.6x150mm) | UPLC (BEH C18, 1.7μm, 2.1x100mm) |
|-----------|------|------|
| Run time | 32 min | 11 min |
| Sema retention time | 18.4 min | 7.2 min |
| Peaks detected (>0.05% area) | 6 | 9 |
| Reported purity | 98.7% | 97.4% |
| Main impurity peak | 0.8% at 16.1 min | 0.6% at 6.3 min + 0.5% at 6.8 min |
The UPLC resolved what appeared as a single 0.8% impurity on HPLC into two separate peaks (0.6% + 0.5%). It also detected three additional minor impurities between 0.05-0.15% that the HPLC simply couldn't resolve.
Does this mean the HPLC result is "wrong"? No. It means HPLC gives you a less detailed picture. For most practical purposes — deciding whether a product is acceptable quality — either method works. A 98.7% purity by HPLC for semaglutide is good. A 97.4% by UPLC for the same sample is also good. You just can't directly compare the two numbers without context.
My recommendation: look at the method details on the COA. If a lab reports ">99% purity by HPLC" using a short column with a fast gradient, be slightly skeptical. Their method might lack the resolving power to detect impurities.